I doubt that the calibration fluid is the problem - we had a few bottles with 1.4mS in it instead of 50mS, if this would be the case he would measure complete different values, but these values are nearly OK
the influence of the temperature compensation is very high for conductivity measurement, that means that the temperature of the calibration fluid must be entered very accurately
just for testing - to check the probe and hardware with eliminating the temperature influence you can do this:
calibrate and enter a fix temperature, e.g. 25°C, and set a fixed temperature for the compensation instead of using the temperature probe
-> after this the ProfiLux should show a reading of 49.8 ... 50.2 mS when immersing the probe into the calibration fluid right after the calibration process