    Ausrufezeichen PH reading differencies

    Hi,

    I have an alarming situation with my PH readings through Profilux and ProfiluxControl -program. ProfiluxControl is showing PH as 10,5 (when connected with usb cable to pc) and profilux is showing 5,15 (when not connected to pc). I have tested the PH-probe in RO/DI water and the reading was around 7,4 (I saw the figure change, when connected to pc). The PH-probe has been calibrated many times and still the same problem continues. Also, when displaying the measurement data, it shows strange staright line from the days I have not connected the profilux to my pc... Something must be off and not right. What to do?

    ProfiluxControl 5.1.7.2
    Profilux firmware 5.17

    measurements2.JPG

    measurements1.JPG

  2. #2
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    there are 2 possibilities:

    a) the probe is bad

    b) there is a problem with a ground loop (another device in the tank causes interferences, and they are transmitted this path: water-probe-ProfiLux-USB-cable-laptop
    -> find device
    or maybe a grounded power supply for ProfiLux (with earth pole) will help

    did it ever work?
    if so - what caused the change?
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  3. #3
    Join Date
    09.05.2014
    Location
    Helsinki
    Posts
    8

    Default

    Hi Matthias,

    Thank you for your reply.

    I changed the Profilux wall plug to a different socket (I had all my tank electronics connected to the same socket) and that changed the PH figure to 8,2 in Profilux (when not connected to the pc). Now that is the correct figure, i am after. However, again, when I connected the Profilux to my pc with the usb cable, I am getting a different reading (9,2). So there must be something causing troubles with the connection to pc? Can it be the cable or the software? Or is my pc causing the interference? This is very odd...

  4. #4
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    Hello,

    this could be one reason, see Matthias' comment:

    "b) there is a problem with a ground loop (another device in the tank causes interferences, and they are transmitted this path: water-probe-ProfiLux-USB-cable-laptop
    -> find device
    or maybe a grounded power supply for ProfiLux (with earth pole) will help"
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

Tags for this Thread

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •