    Hi

    I just calibrated my conductivity probe using the 50ms solution.

    All went well but when i put it back in the solution at the same temp as the tank, I got a 49ms reading. I thought that was quite close but decided to calibrate again to see if i could get it more accurate.

    I followed the same procedure again (but it had been in rodi water for about 5 mins) and this time the solution read 32.9ms!

    I did it one more time and it still read 32.9ms. (wish i had left it alone!)

    What have i done wrong? and how can i fix this problem?

    thanks

  2. #2
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    You should simply repeat the calibration procedure.

  3. #3
    Join Date
    21.11.2011
    Location
    manchester
    Posts
    108

    Default

    I have repeated the calibration process 4 times now and the reading of the fluid still comes out at 32.9-32.8

    I have now ran out of calibration fluid.

    could the probe be defective?

  4. #4
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    The first calibration went ok, didn't it?

    Did you change anything before the second try?
    Could you post the ADC (ADC1 / ADC2) values for the conductivity probe reported in ProfiLuxControl?
    If you are using salt water with a ProfiLux3eX or ProfiLuxPlus2eX the range should set to salt water. If ProfiLuxControl shows Conductivity(S), the range setting seems ok.

  5. #5
    Join Date
    21.11.2011
    Location
    manchester
    Posts
    108

    Default

    Yes the first one came out at 49 with the probe in the solution after the calibration

    I left the probe in ro water and tried again

    the ADC values were pretty constant during calibration with ADC1 at 19-20 and ADC2 at 449-451

    The probe is set to sea water in profilux 3ex

  6. #6
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    Hm, the ADC values seems good - the higer value is perhaps a little low - but I have no reverence value now.

    Which firmware version is running and which model do you have?
    Did you a firmware update?

    Maybe a solution:
    save your settings and sensor data via PLC
    make a factory reset on your ProfiLux
    load the settings to the ProfiLux via PLC
    check the ProfiLux settings
    retry the calibration (you could reuse the calibration fluid for those tests - if the calibration went ok you could repeat it with fresh calibration fluid)

  7. #7
    Join Date
    11.01.2011
    Posts
    28

    Default

    I'm having the same problem. Readings of 32.8 after calibration.

    sent from the bottom of the sea using tapatalk

  8. #8
    Join Date
    21.11.2011
    Location
    manchester
    Posts
    108

    Default

    Just wanted to add that this problem was me being a total dumb ass!

    You have probably got your reading set to the wrong type of measurement. i cant remember off hand what it was but when i changed the measurement reading type it turned out the probe was calibrated correctly!

    doh!

  9. #9
    Join Date
    11.01.2011
    Posts
    28

    Default

    Fixed it. Very odd. Switch from salt, to fresh, and back. Now it works.