Thread: Cond(s)/pH extension - which port is which?

  1. #1
    Join Date
    28.05.2009
    Posts
    23

    Default Cond(s)/pH extension - which port is which?

    Hopefully a simnple question - as per subject heading I'm wondering which port in my Cond(s)/pH extension card is which? They both take bnc fittings. I took my housing apart as I upgraded my IIex to a IIIex and when trying to calibrate one of my pH probes it only reads 10.5 no matter how many times I recalibrate in the lower port and I don't want to try the upper port in case that can damage it (I was "fairly" sure I had it in the right one).
    Any advice?
    thanks
    Mark

  2. #2
    Join Date
    19.02.2009
    Posts
    2,976

    Default

    Hello,
    here an excerpt of the instruction that should come with the extension card:
    "The upper, back standing BNC plug is used to connect the pH electrode, the lower, out sticking BNC plug is used to connect the conductivity electrode. Please ensure not to swap these connectors!"

  3. #3
    Join Date
    28.05.2009
    Posts
    23

    Default

    Thanks - fixed!

  4. #4
    Join Date
    03.03.2009
    Posts
    81

    Default

    What happens if they are swapped and calibrated?

  5. #5
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    Quote Originally Posted by freemchr View Post
    What happens if they are swapped and calibrated?
    wrong measurements, but nothing will be damaged