    I just tried a temperature check of my sensors (including the GHL one) with ro water method with ice bath (https://www.nist.gov/sites/default/f...ions/sp819.pdf) and all mercury is getting 0C, but the GHL sensor is 0.6C too high (ADC 677 / 319 calibration data) - is it worth calibrating the GHL probe regulary?

  2. #2
    Join Date
    27.05.2015
    Posts
    2,492

    Default

    Calibration for this probe is not necessary, but can be adjusted if needed.

    Go to the temperature settings page and you will see the option to adjust the temperature readings. If you don't see the option, make sure you have EXPERT MODE enabled, located in the EXTRAS tab at the top of the screen.

  3. #3
    Join Date
    07.04.2017
    Posts
    157

    Default

    Quote Originally Posted by Vinny View Post
    Calibration for this probe is not necessary, but can be adjusted if needed.

    Go to the temperature settings page and you will see the option to adjust the temperature readings. If you don't see the option, make sure you have EXPERT MODE enabled, located in the EXTRAS tab at the top of the screen.
    Will the adjustment overwrite the factory calibration data or is it just an offset of those data?