    Default cant calibrate in GCC

    Why cant I calibrate my sensors in GCC? There is no tab or option to do so. I can do it via the touch screen...
    When calibrating pH via the touch screen it tells me to use a calibration fluid w a pH value I don't have.
    what gives?

  2. #2
    Join Date
    27.05.2015
    Posts
    2,488

    Default

    Calibration is carried out on the ProfiLux itself.

    1. Press any of the arrow keys on the controller.
    2. Select SENSOR SETTINGS.
    3. Select the sensor you wish to calibrate.
    4. Select CALIBRATION.
    5. Follow on - screen commands

    If this is a new probe, calibration tolerance should be set to 1.

    You will be able to adjust the calibration fluid values during the calibration setup steps.