    Default Redox probe problems

    My second redox probe is stuck at 128. Probe is maybe 4 months old. Just started a week ago.

    Calibrated three separate times and it goes back down to 128.
    Doesn't fluctuate at all during the day

    I know it's not right because other redox probe I have is reading at 250-300 daily average

    Any ideas?

    Running latest FW versions


    Always at sea;

    Dr. Antonio Longo

  2. #2
    Join Date
    27.05.2015
    Posts
    2,492

    Default

    Antonio,

    Have you tried switching the Redox probes and seeing if the problem carries over?

  3. #3
    Join Date
    21.06.2011
    Location
    South Florida
    Posts
    827

    Default

    Yes I have.
    No changes


    Always at sea;

    Dr. Antonio Longo

  4. #4
    Join Date
    27.05.2015
    Posts
    2,492

    Default

    Can you please post screenshots of each Redox probe calibration data?

  5. #5
    Join Date
    21.06.2011
    Location
    South Florida
    Posts
    827

    Default










    Always at sea;

    Dr. Antonio Longo

  6. #6
    Join Date
    18.02.2009
    Location
    Kaiserslautern
    Posts
    7,710

    Default

    did I understand correctly - swapping the probes sho that both probes are OK but always the same sensor port will not do anything?

    this would mean that the port died (for whatever reason) - but: how is it posible then that you were able to calibrate the probe?

    what value does ProfiLux show if youput the redox probe into the calibration fluid? is 220mV displayed? or no reaction here, too?
    No support or warranty issues over PM! Please send PMs to the moderators only if you have general problems with using the forum! Thanks for helping us to keep the support efficient.
    Kein Support oder Reklamationsabwicklung über PM! Bitte senden Sie an die Moderatoren nur PMs bei allgemeinen Problemen mit der Verwendung des Forums! Danke dass Sie uns dabei helfen, den Support effektiv zu gestalten.

  7. #7
    Join Date
    21.06.2011
    Location
    South Florida
    Posts
    827

    Default

    I can calibrate the probe. 220 mv displayed after calibration
    But when I put it back in the tank 12 inches away form redox probe 1, it stays around 129-130


    Always at sea;

    Dr. Antonio Longo

  8. #8
    Join Date
    23.05.2010
    Location
    aalborg
    Posts
    173

    Default

    i once had this problem, due to electronic interfering..