Not sure what is happening but the KH Director last measurement taken does not match on when I either timed it to happen or when I manually started the measurement. My time is local PC and UTC is +0. I am in Wash DC so should set my UTC to -5?
Printable View
Not sure what is happening but the KH Director last measurement taken does not match on when I either timed it to happen or when I manually started the measurement. My time is local PC and UTC is +0. I am in Wash DC so should set my UTC to -5?
GMT is 0, Washington is 5 time zones before Greenwhich, so I think UTC -5 is correct for you.
Shouldn't my KH measurement time match local pc time and not the UTC?
all times are stored in UTC
but shown is the time according to the time zone you selected for your device
you wrotein your first postQuote:
UTC is +0
did you change the ProfiLux setting meanwhile to your correct timezone?
Was this ever figured out? I'm in EST (-5) as well and have my p4 kh director set to reflect that (-5). It worked for awhile, but now shows it 4 hours behind of when it's actually taken. Even if I manually test
Hi there,
There was nothing to figure out. It's a matter of changing the time zone setting in your P4 to match your timezone. :)
This is a separate setting from the actual internal clock on the P4. The timezone settings are used by your KHD and probes. You will find this in the GENERAL settings page.
Check the settings again and post a screenshot if you're still having trouble with it.
Oh ok, mine is set to correct time zone, but only worked for a week showing the correct time. Then it went to 4 hrs behind. Must be something else